I agree, if you change to the Vocarb 3000 trap you can desorb @ 250 and bake @260.
It helps but I have had the same carryover problems for years while using the Vocarb 3000 with the higher temperatures.
How bad is the carryover? Is it 1ppb after a 200ppb standard, or is it 5ppb after a 50ppb standard?
The only systems I have that I have gotten rid of this problem on are my newer ones with the Centurion autosampler where I can wash the sparge tube 3 times with 90C water then bake the sparge tube at 90C for the remainder of the bake time while also rinsing the sampling needle and lines of the autosampler with 90C water for 1 minute and flushing with He for another minute. Anything less and I will get carryover for these compounds at 1ppb after a 25ppb standard even on instruments that only run drinking water samples.
Another place that seems to contribute to the carryover problem is the moisture reduction systems. I found that if you bake at a lower temperature you get more time before needing to clean or replace the moisture trap tube. It seems the high temperature bake pyrolizes trace amounts of organics to form what I think is a char inside the tubes that will work as a trap later on(just a theory I haven't been able to prove it yet). If I bake at about 180C I get longer life from the moisture trap tubes before the carryover happens. If the tubes are clean, you shouldn't need to go any higher with the bake to remove the traces of the target analytes.
As mentioned above, heating the sparge tube between samples seems to give the lowest carryover of anything I have tried so far.