I bought a new molecular sieve plot column (HP) for analysis of sub-ppm amount of oxygen and argon, I am sure the column could seperate them, but, at trace level, even tens ppm of oxygen could not be detected.

My question is, can I use air or oxygen as carrier gas to regenerate this column?
No :shock:
A molsieve column is just holes of a certain size eg 5A. If you ran air through it the holes would be full of CO2 that sticks to a Molsieve. Regeneration is done by heating it up and leave it purging with carrier gas. What are you running ? Helium , Nitrogen ?

The separation of Argon, Nitrogen and Oxygen is tough, it is most likely that currently they co-elute.
I know that if oxygen peak decrease happened on MS packed column can be resolved by regeneration with oxygen or clean compressed air as carrier gas at high temperature, but what about MS PLOT columns, do they contain some kinds of organic binder or something which can be damaged by oxygen?
3 posts Page 1 of 1

Who is online

In total there are 151 users online :: 0 registered, 0 hidden and 151 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 151 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry