Could you please advise me for the following:
Two weeks ago I cleaned the Ion source (MSD 5975) and Autotune was ОК. After 40 samples on HSP, the calibration was perfect too.
But the problem has appeared:
Autotune goes bad, there is some moving of mass, mass 502 moved at 504. When the Autotune was repeated the error “Mass gain – out of range” occurred.
What are the possible reasons for the movement of big masses?
How to proceed for resolving the problem – put the masses in their range?
I can clean the Ion Source again, but it’s not normal so frequent and only after 40 samples to do this and it could not resolve the problem for a long time.
Please note that there is sufficient quantity of calibrating agent into “Calibration vial”.
The image applied here is from Tune evaluation records:

Looking forward to your comments.