Hi
1-I use capric acid (ester)as internal standard in analysis of methyl oleate.
but when the area of analyte increase, area of ISt deacrease or conversely.
I haven't any explanation for it.

2-how do I calculate the response factor?(is it correct: area/amount of analyte)in quantitation analysis,should use the response factor?
please explain for me
thanks in advance
Your results do not appear authentic. I would suggest that you explain your procedure in more detail. Perhaps then a reasonable explanation for your unusual results can be postulated.
Hi
internal standard:methyl decanoate
analyte:methyl oleate
column:PEG 20M capillary column,30m,0.22mm id,
carrier gas and make up:He 1.53mm/min
hydrogen and air:30,300 mm/min
det temp(FID):250
inj temp:240
column temp:160,5C/min,230
split ratio:1:50
conc istd:50ppm
thanks

You do normal external standard calculations EXCEPT that you use the ratio of your analyte/internal std peaks instead of your analyte area everywhere in the calculations.
Thanks,
DR
Image

thanks alot
nina
5 posts Page 1 of 1

Who is online

In total there are 44 users online :: 1 registered, 0 hidden and 43 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 43 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry