Any way to enhance molecular ion for GCMSMS analysis?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
I am just starting to develop some methods for GCMSMS and am wondering if there is a way to get less initial breakdown of the analyte in the EI source so that you have greater molecular ion to work with going into the collision cell.

I have tried lowering the uamp current to the filaments and am starting to play with reducing from -70ev but not really seeing any shift in abundances. Is there any trick to this or am I just stuck with a lot of fragmentation in the source?
The past is there to guide us into the future, not to dwell in.
Cooling the source might help a little or what about CI?
Cooling the source might help a little or what about CI?
Well I pushed for CI when ordering the instrument, but ended up getting one without it. Now I wish we had.

Working on a method for Methamphetamine without derivatization, seems I can get a little better sensitivity in SIM than I can with MRM, just more background is present. The molecular ion is probably less than 1% abundance so the molecule breaks down pretty well before hitting the collision cell.
The past is there to guide us into the future, not to dwell in.
James,

How low have you dropped from 70 eV ?? You may have to go quite low, e.g 20 eV to get a meaningful mol. ion. If you have a solids probe, this would be easy to do stepwise.

Since your M+. is at m/z 149, and the fragments are obviously lower, you could distort the tuning (I assume QQQ) to enhance the low mass end; you don't need sensitivity above ~m/z 175 in either Q1 or Q3.

JMB
James,

How low have you dropped from 70 eV ?? You may have to go quite low, e.g 20 eV to get a meaningful mol. ion. If you have a solids probe, this would be easy to do stepwise.

Since your M+. is at m/z 149, and the fragments are obviously lower, you could distort the tuning (I assume QQQ) to enhance the low mass end; you don't need sensitivity above ~m/z 175 in either Q1 or Q3.

JMB
I think I went down to 30ev, I may try lower and see what happens.
The past is there to guide us into the future, not to dwell in.
5 posts Page 1 of 1

Who is online

In total there are 6 users online :: 0 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry