I´m having a problema with my RID detector. My system is an Agilent 1100. The problem is the baseline. Waves. Different mobile phases and columns. Purged the references several times, prepared and degassed the mobile phases more than once, the temperature is stable, room and in the detector. The first problem was, ACN:Water 1:1, flow 0.6 mL/min, detector and column T 50 °C, C18. Pumping with channel B. The last thing I tried was to change to channel A. Problem solved. Next problem. MeOH:Phosphate buffer (1:4), flow 1.2 mL/min, C18, T 35 °C in column and detector. Channel C. When changed to channel A, problem solved. But, before me, some peolpe had problems and they were using channel A.
I cannot realise if it is the RID or the pump, perhaps the mixing valve or some leak I´m not seeing. When I turn the pump to 0 mL/min, the baseline becomes stable, so I think it´s not a temperatur problem .
Any ideas?