In our lab, we run several analysis in which the aqueous phase has pH around 1.5-1.7. In some the organic phase is 20% and in others is 80%. How can i be sure of the final pH of my mobile phase so to protect my columns?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
If you are running existing validated methods, you run them as written, period. Column lifetime is a secondary issue.
If you are developing methods, the issue is complicated. Read these three articles by Bill Tindall that originally appeared in LC-GC:
http://www.lcresources.com/tsbible/20112002.pdf
http://www.lcresources.com/tsbible/20122002.pdf
http://www.lcresources.com/tsbible/21012003.pdf
If you are developing methods, the issue is complicated. Read these three articles by Bill Tindall that originally appeared in LC-GC:
http://www.lcresources.com/tsbible/20112002.pdf
http://www.lcresources.com/tsbible/20122002.pdf
http://www.lcresources.com/tsbible/21012003.pdf
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 46
- Joined: Mon Feb 11, 2013 10:13 pm
nice articles(by the way there were 2 of them) but i didn't find a solution. How you chromatographers handle this issue?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Oops! I copied the second link twice.
I've edited the post so all three links are correct.
That said, the answer is in the first article:

That said, the answer is in the first article:
So, the short answer to your question is "you can't". To reiterate: if you're running an existing, validated method, you run it as written, period. If you're developing a method, you try to stay within the pH limits specified by the manufacturer for that column, recognizing that the lifetime will be shorter at the extremes of those limits than it will in the middle. My mantra on that is "Columns are cheap; time is valuable."However, if I knew an acid sample was 50% dissociated at aqueous pH 5, I have no way to predict what the dissociation will be in methanol–water at pH 7. I don’t know the pKa of the acid in the mixed solvent, and, for the reasons described above, I can’t calculate hydrogen ion activity from the pH measurement. Furthermore, if I knew the sample or column was unstable at aqueous pH 7, I could not predict the stability in the pH 7 mixed solvent because I don’t know the hydrogen ion activity or the effect of the standard state change. By itself, neither the pH 5 measurement of the aqueous buffer nor the pH 7 measurement in the mixed solvent is much use in predicting retention and stability behavior in the pH 7 mobile phase.

-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 1888
- Joined: Fri Aug 08, 2008 11:54 am
... and of course the column manufacturers are in exactly the same position. So if they say that a column is stable at pH 9.0, they mean that it is safe to include aqueous buffers at pH 9.0 in any appropriate solvent mix you make up for use in the column.
-
- Posts: 421
- Joined: Wed Jul 18, 2012 6:45 am
You still can use the good old fashioned pH indicator paper. Prepare your mobile phase and put some drops on an indicatior paper and see the magic. It is not as accurate as a pH-meter, but it will give you an idea. The dyes are in the same situation as your stationary phase is: They have to decide if they want to protonated or not in presence of organic modifier.
-
- Posts: 512
- Joined: Tue Jul 19, 2011 9:26 pm
Just two throw in my 2 cents: don't concentrate to much on the (apparent) pH of the mobile phase, also keep an eye on other parameters of the method. Type of buffer, concentration, aqueous/organic ratio and especially temperature all play their role concerning column stability. A phosphate buffer pH 9 can be more demanding for a column than a pyrrolidine buffer pH 10.5
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
My 2 cents: use the pH meter to figure out how much modifier you need to add by either weight or volume to reach that pH needed for the assay, then use one of those two techniques to make mobile phase in the future (document this). Those other ways are much more accurate.
8 posts
Page 1 of 1
Who is online
In total there are 26 users online :: 1 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google Adsense [Bot] and 25 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google Adsense [Bot] and 25 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science