Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
I have a HP 6890 Series GC set up with MS and FID. When I control the instrument from the instrument panel, I am able to have dual injections and run both columns at the same time. However, when I try to run the instrument through chemstation (G1701BA), it will not let me run both columns simultaneously.

Has anyone encountered this issue before? Any suggestions?

Thanks!
is your sequence configured correctly to run simultaneous injections? its a little tricky but the help documentation explains how to properly set it.

if i remember correctly you need to invoke the unlinked keyword in the first line, and alternate lines for each injection (front and back) to specify file names. check out the documentation for details.
Yes, there are various combinations of keywords in the sequence - unfortunately this has varied over the MSD chemstations. Look for help on stagger, unlinked, and makesec keywords in the help.
3 posts Page 1 of 1

Who is online

In total there are 44 users online :: 0 registered, 0 hidden and 44 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 44 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry