FID setpoints greyed out, Chemstation 5.04, 6890+ GC
Posted: Wed Oct 08, 2014 6:45 pm
A 6890 Plus GC in a teaching lab is used twice a year for only a few samples every 6 months.
This year when we started up the instrument we got the introductory screen as if the Chemstation software (v5.04) had been newly updated (What's new and Tutorial popup). We were immediately suspicious since it hadn't been touched since the spring.
Problems that immediately showed up included not being able to edit or save any GC methods. This even though the the Chemstation access level had not been set to Operator (and no Manager access login screen showed up when Chemstation started up.
This problem was ultimately solved with a tip here in the Chromatography Forum (thanks): we edited the win.ini file to delete the "LEVEL=" line (we DID find an entry in that file "LEVEL=SOMETHING" etc). No clue how that came about since the instrument worked perfectly in the spring. (No angry chemistry students at our university...)
OK so now we are back to being able to edit and save our methods again.
But all is not OK in the Instrument | Edit| Detectors dialog for the 6890 FID:
We can light the FID by hand on the GC.
The FID's signal is being seen/recorded by Chemstation.
We can change the FID's heater temp setting (and click box) and its actual temp changes are reported in Chemstation and on the GC.
We can change FID Lit Offset value and that change is reflected in the software and on the GC.
We can turn the flame off and on in the software via flame and click box + Apply. Flame's actual toggles "on" to "off" in Chemstation as expected.
However, ALL the other software FID detector ACTUALS and SETPOINTS are greyed-out and therefore can't be edited (even though their click boxes work and those changes are reflected on the GC front panel.)
That is, the H2, air, and makeup setpoints can NOT be set in the software and when the flame is lit and H2, air and make gases are clearly flowing, the actuals are greyed-out and aren't reported in the software. Therefore we have no control on the H2, makeup, and air flows. Unlike our 7890A, the GC's FID flows CAN'T be set on the GC's front panel so we're in a fix.
No hardware, detector, cabling, or gas line changes have been modified on this instrument in many many years.
It seems that the chance that hardware problems initiated this problem is small.
We still don't know how the Manager access password got set but that does seems to be the genesis of the problems.
Reinstallation of Chemstation 5.02 and then upgrade to 5.04 (with a floppy disk!) did not affect/solve the problem. The "Level=" win.ini code line did not reappear after reinstallation so apparently the default setting is Manager, unpassworded,as we had experienced with this software for years.
Just to make this harder, the Chemstation logbook shows no errors on startup.
Otherwise the Chemstation software runs flawlessly, and of course we boot the GC before we boot Chemstation to avoid the "68GC is missing/not configured" error message.
Suggestions as to how we can gain control of our FID gas flows again would be greatly appreciated.
Thanks for your time in advance.
Red Head
This year when we started up the instrument we got the introductory screen as if the Chemstation software (v5.04) had been newly updated (What's new and Tutorial popup). We were immediately suspicious since it hadn't been touched since the spring.
Problems that immediately showed up included not being able to edit or save any GC methods. This even though the the Chemstation access level had not been set to Operator (and no Manager access login screen showed up when Chemstation started up.
This problem was ultimately solved with a tip here in the Chromatography Forum (thanks): we edited the win.ini file to delete the "LEVEL=" line (we DID find an entry in that file "LEVEL=SOMETHING" etc). No clue how that came about since the instrument worked perfectly in the spring. (No angry chemistry students at our university...)
OK so now we are back to being able to edit and save our methods again.
But all is not OK in the Instrument | Edit| Detectors dialog for the 6890 FID:
We can light the FID by hand on the GC.
The FID's signal is being seen/recorded by Chemstation.
We can change the FID's heater temp setting (and click box) and its actual temp changes are reported in Chemstation and on the GC.
We can change FID Lit Offset value and that change is reflected in the software and on the GC.
We can turn the flame off and on in the software via flame and click box + Apply. Flame's actual toggles "on" to "off" in Chemstation as expected.
However, ALL the other software FID detector ACTUALS and SETPOINTS are greyed-out and therefore can't be edited (even though their click boxes work and those changes are reflected on the GC front panel.)
That is, the H2, air, and makeup setpoints can NOT be set in the software and when the flame is lit and H2, air and make gases are clearly flowing, the actuals are greyed-out and aren't reported in the software. Therefore we have no control on the H2, makeup, and air flows. Unlike our 7890A, the GC's FID flows CAN'T be set on the GC's front panel so we're in a fix.
No hardware, detector, cabling, or gas line changes have been modified on this instrument in many many years.
It seems that the chance that hardware problems initiated this problem is small.
We still don't know how the Manager access password got set but that does seems to be the genesis of the problems.
Reinstallation of Chemstation 5.02 and then upgrade to 5.04 (with a floppy disk!) did not affect/solve the problem. The "Level=" win.ini code line did not reappear after reinstallation so apparently the default setting is Manager, unpassworded,as we had experienced with this software for years.
Just to make this harder, the Chemstation logbook shows no errors on startup.
Otherwise the Chemstation software runs flawlessly, and of course we boot the GC before we boot Chemstation to avoid the "68GC is missing/not configured" error message.
Suggestions as to how we can gain control of our FID gas flows again would be greatly appreciated.
Thanks for your time in advance.
Red Head