gcms signal consistently decrease within the assay
Posted: Mon Sep 29, 2014 3:23 pm
by jiangds06
I saw signal decrease after each run. Sample preparation is liquid liquid extraction with ethyl acetate. Is this because of residual water in the ethyl acetate phase?
Re: gcms signal consistently decrease within the assay
Posted: Mon Sep 29, 2014 6:34 pm
by Steve Reimer
Increasing water in the MS source will cause a loss of response.
Re: gcms signal consistently decrease within the assay
Posted: Tue Sep 30, 2014 1:00 pm
by jiangds06
Thank you so much. So how can fix this issue?
Increasing water in the MS source will cause a loss of response.
Re: gcms signal consistently decrease within the assay
Posted: Tue Sep 30, 2014 2:17 pm
by bunnahabhain
Use anhydrous sodium sulfate to dry your ethyl acetate sample after extraction.
Re: gcms signal consistently decrease within the assay
Posted: Thu Oct 02, 2014 5:12 pm
by jiangds06
tried that. but analytes are adsorbed in the sodium sulfate.
Use anhydrous sodium sulfate to dry your ethyl acetate sample after extraction.