- Posts: 3
- Joined: Thu Sep 18, 2014 4:17 pm
Some background: Im using a buffer with sodium dihyrogen orthophosphate (ph to 3.1 with ortho acid), my needle washes are varients of Organic and Water. Its not a great method but it has worked fine in the past, my issue is the second half of my 4 minute injection gets very noisy and the peak tends to either split when it comes off with the background noise, or else have massive tailing. The first half of the injection is fine, its just goes very wrong during the second half. I have checked and triple checked the method so my gradient and wavelength is right. I have swapped over mobile, washes, and columns from a back up machine which is working fine but the problem was still there. The flow cell was washed out with water, acid and flushed again with organic, the problem cleaned up a bit but not to the extent i need it to. Is there anything else that can be done to improve the baseline. Note there are no leaks on system.
The only thing is that I usually dont clean out the columns with organic etc before use, i just start using it straight way could this be a factor?














 
																							