captopril tablets USP method
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
has anybody faced problem with analysis of captopril tablets using USP method
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Don't know the method, but if you are getting strange peak shapes with captopril you are probably at the wrong pH. Can you describe the method and the problem?
-
- Posts: 6
- Joined: Wed Aug 31, 2005 3:34 pm
the dissolution method invovles
media: 0.01nHcl, 900ml
basket:50rpm
detection by hplc
mobile phase: buffer: MEOH : 450: 550
buffer = (0.5ml OPA in 450ml WATER)
column :30cm c18 with 10microns
wavelength :220nm
USP suggest to inject samples immediately before 8hrs
i usually encountered problems like high Dissolution, and samples have to be injected within 2hrs.
secondaly assay with same chromatographic setup but diluent being diff
shows 100% assay if injected immediately and higher assay after even 1hr.
[/u]
media: 0.01nHcl, 900ml
basket:50rpm
detection by hplc
mobile phase: buffer: MEOH : 450: 550
buffer = (0.5ml OPA in 450ml WATER)
column :30cm c18 with 10microns
wavelength :220nm
USP suggest to inject samples immediately before 8hrs
i usually encountered problems like high Dissolution, and samples have to be injected within 2hrs.
secondaly assay with same chromatographic setup but diluent being diff
shows 100% assay if injected immediately and higher assay after even 1hr.
[/u]
-
- Posts: 392
- Joined: Thu Sep 08, 2005 4:12 am
when you are talking about "USP suggest to inject samples immediately before 8hrs ; i usually encountered problems like high Dissolution, and samples have to be injected within 2hrs.
secondaly assay with same chromatographic setup but diluent being diff
shows 100% assay if injected immediately and higher assay after even 1hr. " are you meaing that the detected amount increases with the time of sample storage before injection such as 40% at 10:00 AM when the sample is prepared and injected, 45% at 11:00 AM if the same sample was injected again, and 50% at 12:00 AM if the same sample was injected again....?
If so, could you set up a way so you always inject the samples with the same delay or immediately? I am sorry to bother you.
secondaly assay with same chromatographic setup but diluent being diff
shows 100% assay if injected immediately and higher assay after even 1hr. " are you meaing that the detected amount increases with the time of sample storage before injection such as 40% at 10:00 AM when the sample is prepared and injected, 45% at 11:00 AM if the same sample was injected again, and 50% at 12:00 AM if the same sample was injected again....?
If so, could you set up a way so you always inject the samples with the same delay or immediately? I am sorry to bother you.
Excel
-
- Posts: 480
- Joined: Fri Sep 03, 2004 7:57 am
I think Captopril is unstable in aqueous media. You may store the sample in cooler, in vials those closed tightly to reduce degradation rate. Is it light sensitive too?
5 posts
Page 1 of 1
Who is online
In total there are 20 users online :: 4 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Google [Bot], jpappas_PNB, Semrush [Bot] and 16 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Google [Bot], jpappas_PNB, Semrush [Bot] and 16 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science