SH-GC-MSD Residual ethanol
Posted: Wed Sep 24, 2014 5:08 pm
Thanks for all your advice regarding static headspace which is new to me. I've been reviewing the forum archives especially for Chramatographer1's posts and all the residual ethanol posts and HS parameters.
Here is what I've come up for for my residual ethanol method. I get two types of sample liquid flavors in soybean oil, propylene glycol, triacetin, or Neobee and spray dry encapsulated or pan dry encapsulated. I need to ensure they have minimal ethanol for halaal and my goal is to get down to below 100ppm in the samples for LOQ. I have a Tekmar 7000 deactivated sample pathway (unfortunately except for the cut end of the inlet carrier line) and a 6890N-5975MSD. Liner 1mmID. column Innowax or 1701 30m .25mm .25um film.
Here is what I am thinking. I've read Chromatographer1's advice regarding using a low sample volume in the vial. For diluent I am thinking of Using alpha-pinene. I have a ton available it has good oil and alcohol solubility and some/minimal water solubility with a Bp of 155 deg C. I might need a separate method with water for the encapsulated samples though.
Parameters
-100mg sample weighed into a headspace vial
-300ul of alpha pinene containing standard for standard addition
-9mL vials
Vial pressure 15PSI N2 supplied by 6890 aux EPC
Platen temp 80 deg
Loop temp 165 deg
Tline temp 175 deg
Sample equil 2 min
mix time 8 min
mix power 4 (max recommended for 9ml vial)
post mix equil 1 min
Pres time 0.2min
Pres equil .05min
Loop fil (1 ml) 0.25min
Loop equil 0.05min
inject 1 min
Inlet 200 deg C
3:1 pulsed split 3 ml column 9 ml split 2-3 septa purge so a total of ~15 ml/min transferline for 1 minute
Column flow 1ml/min after 1 minute
40 deg for 5 minutes +3 to 75 deg C +20 to 200 5min.
GC/MS SIM 45&46 m/z
I may have to loose the pulse if ethanol doesn't focus well.
Any comments?
Parameters
-9mL vials
Platen temp 80 deg
Loop temp 165 deg
Tline temp 175 deg
Sample equil 2 min
mix time 8 min
Pres time 0.2min
Pres equil .05min
Loop equil 0.05min
inject 1 min
Inlet 200 deg C
GC/MS SIM 45&46 m/z
Any comments?