autotune
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
9 posts
Page 1 of 1
-
- Posts: 155
- Joined: Tue Sep 27, 2011 8:54 am
I would say that there is a leak somwhere in your system!
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
You have massive leak...check connection between GC and MS
Tomasz Kubowicz
You have massive leak...check connection between GC and MS
Tomasz Kubowicz
-
- Posts: 8
- Joined: Tue May 13, 2014 7:38 pm
checked that and it seems the column is stuck in the GC/MSD interface, the column is not budging
-
- Posts: 67
- Joined: Tue Nov 11, 2008 4:18 pm
Hi
I'd like to remind you that there's a thread about blocked GC/MS transfer lines in the forum. Good luck
I'd like to remind you that there's a thread about blocked GC/MS transfer lines in the forum. Good luck
Mike
-
- Posts: 8
- Joined: Tue May 13, 2014 7:38 pm
well it took some effort but we got the column out...now the tune shows the emvolts higher at 2894 and N2 is 51% O2 16% Co2 2% H2O 12% N2/H2O 418%
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
You still have huge leak. Check GC:
-inlet (do leak test)
-check your carrier gas trap ! (it could be saturated)
You can make simple test to narrow problem down:
-plug transfer line with blank nut and bake out MSD for couple of hours. Than perform autotune.
Regards
Tomasz Kubowicz
You still have huge leak. Check GC:
-inlet (do leak test)
-check your carrier gas trap ! (it could be saturated)
You can make simple test to narrow problem down:
-plug transfer line with blank nut and bake out MSD for couple of hours. Than perform autotune.
Regards
Tomasz Kubowicz
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
Start by loading an older tune file before you run autotune, if you have saved any of the recent ones. Otherwise it may take a long time for the routine to converge on usable values.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Definitely try this, it will tell you if you have a leak in the GC or the MS. You could have lint or some other object on the O-Ring that seals the analyzer side plate or even a broken interface if you find that the leak is still there when you use a solid ferrule and nut on the interface.Hello
You still have huge leak. Check GC:
-inlet (do leak test)
-check your carrier gas trap ! (it could be saturated)
You can make simple test to narrow problem down:
-plug transfer line with blank nut and bake out MSD for couple of hours. Than perform autotune.
Regards
Tomasz Kubowicz
The past is there to guide us into the future, not to dwell in.
9 posts
Page 1 of 1
Who is online
In total there are 133 users online :: 0 registered, 0 hidden and 133 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 133 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 133 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science