Does the size of solvent container matter?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
7 posts
Page 1 of 1
Recently I switched my mobile phase container from 4L to 18L one. The solvent levels are almost at the same height, however I observed higher background pressure and faster flow rate. The baseline curves a little bit after each injection. Is it normal to see such change? If the solvent almost used up, will the background pressure lower down and flow rate slower down? If in that case, probably I couldn't get consistent results. If I have to use the 18L solvent container, has it a way to solve it? Thank you very much!
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
Very strange. What flow rate and pressure are you observing? The inlet pressure is usually negligible compared to the pump pressure, and has no measurable effect on retention. Does the larger container have also a larger supply tube to the pump?
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
-
- Posts: 5
- Joined: Fri Aug 19, 2005 4:33 pm
I use the same tubing. But the containor is sit on the bench, at the same height of the pump.
-
- Posts: 5
- Joined: Fri Aug 19, 2005 4:33 pm
my flow rate is normal: 1 mL/min.
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
That is enough mobile phase for 12 days of nonstop running! Are you planning to go on vacation?
I was wondering if perhaps your supply line was restrictive and causing the pump to be inefficient with one reservoir. At 1.0 mL/min there should be no trouble at all.
Does a switch back to the original container also give the original pressure and retention?
I was wondering if perhaps your supply line was restrictive and causing the pump to be inefficient with one reservoir. At 1.0 mL/min there should be no trouble at all.
Does a switch back to the original container also give the original pressure and retention?
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
Maybe someone thought it was an aquarium and put fishes in?
Seriously, what happens if you now place the liquid of the big container into the little one and running your stuff?
Seriously, what happens if you now place the liquid of the big container into the little one and running your stuff?
-
- Posts: 480
- Joined: Fri Sep 03, 2004 7:57 am
Mr. Mueller, I have ever shocked by a lab using huge plastic jug as their mobile phase (especially for buffer solution). I think its inner side will be difficult to be washed or cleaned. Especially if we put fishes in it.
That’s another problem in big container.
That’s another problem in big container.
7 posts
Page 1 of 1
Who is online
In total there are 18 users online :: 3 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Baidu [Spider], Google [Bot] and 15 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Baidu [Spider], Google [Bot] and 15 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science