We have an isocratic method that runs on a 4.6x150 mm Kinetex C18 column (2.6 µm). Flow 1 ml/min, using phosphate buffer/acetonitrile.
Plate count of main peak with this setup is about 20.000 plates. To save some time I turned up the flow to 2 ml/min, and then the plate count was only 5000?? I put the flow back to 1 ml/min and got back my 20.000 plates.
The Van-Deemter plot of the Kinetex material cannot explain this drop (flat line almost at higher velocities). Quite strange?