Posted: Wed Sep 17, 2014 1:16 am
I'm currently developing, and almost at the stage to optimizing method using UPLC. I'm running a gradient programme (ACN:water) with post-time.
Problem faced :
1. Initial pressure at the start of the day is not always the same. (ie last week can be 175Bar, today can be 130Bar and next can be 200Bar). All mobile phase are being filter through glass filter. All lines are purged and equilibrium done before the analysis. But still I face this. How to go about this? This making the retention time shift. Or is this not a big issue as long as I inject my Standard solution every time with sample?
2. Few problems at once : Peak not symmetry, broadening of peak.
But this occurs only after months I run with the same method.
Observed non-symmetry peak (like tailing), tried to prepare sample solution in ACN:water instead of purely ACN. After many times of injections, the non-symmetry seems ok. But still see broadening of peak, double the size before the issue occurred. My point of view, the broadening of peak will influence the resolution of peaks, which I have 2 peaks elute at very close Retention Time. Please advice.
I would like to know also, why I observed sample in ACN gives better signal compared to sample in ACN:water?
3. Volume of injection.
I am expecting double signal height if double volume of injection is being made. This happens to sample in ACN but not to sample in ACN:water. Why is this so?
Owning only 1 column, thus cannot do anything with this parameter. But am getting a new column though.
Please help. Thank you.
Problem faced :
1. Initial pressure at the start of the day is not always the same. (ie last week can be 175Bar, today can be 130Bar and next can be 200Bar). All mobile phase are being filter through glass filter. All lines are purged and equilibrium done before the analysis. But still I face this. How to go about this? This making the retention time shift. Or is this not a big issue as long as I inject my Standard solution every time with sample?
2. Few problems at once : Peak not symmetry, broadening of peak.
But this occurs only after months I run with the same method.
Observed non-symmetry peak (like tailing), tried to prepare sample solution in ACN:water instead of purely ACN. After many times of injections, the non-symmetry seems ok. But still see broadening of peak, double the size before the issue occurred. My point of view, the broadening of peak will influence the resolution of peaks, which I have 2 peaks elute at very close Retention Time. Please advice.
I would like to know also, why I observed sample in ACN gives better signal compared to sample in ACN:water?
3. Volume of injection.
I am expecting double signal height if double volume of injection is being made. This happens to sample in ACN but not to sample in ACN:water. Why is this so?
Owning only 1 column, thus cannot do anything with this parameter. But am getting a new column though.
Please help. Thank you.