drifting baseline + increased solvent front peak
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
We are using a a relatively new column and injecting a pure standard dissolved in ethanol and 2ml of this solution is diluted to 10ml with mobile phase. After 5 injections with the same solution the baseline begins to slope upwards, the standard peak remains sharp and has roughly the same area as the previous peaks. However the solvent front peak (ethanol) increases in size several fold. We have tried cleaning the column and the chromatograms return to normal, but repeat the aforementioned problem after the 5th injection from the same solution. The method is a BP method with a recommended column. Any ideas?
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
1.Mobile phase?
2.Isocratic or gradient?
3.Detector setting? (ref wavelength?)
Tomasz Kubowicz
1.Mobile phase?
2.Isocratic or gradient?
3.Detector setting? (ref wavelength?)
Tomasz Kubowicz
-
- Posts: 3
- Joined: Thu Sep 18, 2014 9:51 am
1) The detector wavelength is 240nm
2) Isocratic
3) mobile phase is a mixture of 6 volumes of 5%w/v of sodium dodecylsulphate in 3%v/v glacial acetic acid, 10 volumes of triethylamine, 20 volumes of butan-2-ol, 310 volumes of acetonitrile and 690 volumes of water. The mixture is then adjusted to pH 3.0 with orthophosphoric acid.
4) The column is a nucleosil C18, 5 micron.
2) Isocratic
3) mobile phase is a mixture of 6 volumes of 5%w/v of sodium dodecylsulphate in 3%v/v glacial acetic acid, 10 volumes of triethylamine, 20 volumes of butan-2-ol, 310 volumes of acetonitrile and 690 volumes of water. The mixture is then adjusted to pH 3.0 with orthophosphoric acid.
4) The column is a nucleosil C18, 5 micron.
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
I would try to inject 5x mobile phase and see if problem still exists. I suspect there is something coming out from column after long time (perhaps something is moving through column and it reaches detector after couple of injections)
Regards
Tomasz Kubowicz
I would try to inject 5x mobile phase and see if problem still exists. I suspect there is something coming out from column after long time (perhaps something is moving through column and it reaches detector after couple of injections)
Regards
Tomasz Kubowicz
-
- Posts: 3
- Joined: Thu Sep 18, 2014 9:51 am
Thanks Tomasz,
The person doing the analysis is putting the column on wash overnight and we will try tomorrow.
The person doing the analysis is putting the column on wash overnight and we will try tomorrow.
5 posts
Page 1 of 1
Who is online
In total there are 51 users online :: 0 registered, 0 hidden and 51 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 51 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 51 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science