I want to calculate the signal-to-noise ratio of a chromatographic peak using Empower software. Is it possible to do it in this software?
Hubert
Advertisement
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
could anyone show the detailed procedure on doing this with the software. I'm just confused because the limit being set for the sensitivity solution i had should be NLT 10 but i'm getting only 0.8.It works well in Empower 2
can you email it to me as well. jayroseville58@yahoo.com thanks
can you send me your detailed explanation on this part? thanks. here is my email ad. jayroseville58@yahoo.comYes it is. If you have the Suitability module it is right in the center of the Suitability tab in the processing method. Be carefully about using the Baseline Noise value though (doesn't work well), and also carful using Detector Noise (RMS noise). The Noise range settings are in the Noise and drift tab for the processing method. Note the Suitability tab even has the option to calculate EP, oh wait now I mean USP noise.
Of course if you don't have Suitability you can make a custom calc to divide peak height by noise (or is that 2x the peak height). Also beware of the scaling factor difference between height and noise, but that value is also available for a custom calc.
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.