fast eluting drug
Posted: Sun Sep 11, 2005 7:02 am
by zydus
dear friends,
i am developing a method for levidopa carbidopa and entacapon combine tablet. the peaks for levidopa and carbidopa are quite close and the resolution is not reproducible all the time. i am using bffer ph3 and acn gradient for 30 min. right now i am trying inertsil ph-3 column. can anybody suggest me how to inprove the resolution and which column is best for non retain compounds.
thanx in advance to all
Posted: Sun Sep 11, 2005 7:19 am
by unmgvar
Go for example at:
www.chrombook.merck.de
look for applications number:
021391
021375
both of them can be twicked a little to be even shorter
hope this helps.
for future needs, ask all your column vendors to provide you with applications CDs. you will find a very broad library which will be very handy in the future.
Re: fast eluting drug
Posted: Sun Sep 11, 2005 5:39 pm
by amaryl
dear friends,
For resolution
try using a column length of 250 mm as it provides a delayed elution and a particle size of 5 micron.
resolution can be achieved by having a step by step increment in ur % B.
by adjustment of flow rates.
by achieveing a balanced wavelength for your two combination of drugs or by wavelength programming.
Posted: Mon Sep 12, 2005 1:05 am
by SIELC_Tech
What kind of retention do you have? What is resolution between and cardidopa. I assume that this two are a critical pair because you should not have a problem retaining entacapone.
Check this method for the retention of DOPA:
http://hplcmethods.com/compound_101.php
Levidopa and carbidopa have different structures, hydrophobicity and pKa you can use these differences in mixed mode chromatography. Retention of DOPAs can be from 3 minutes to over 30 minutes depending on the mobile phase.
For UV you can use sulfuric, phosphoric acids or corresponding buffers
For LC/MS/ELSD you can use ammonium formate, ammonium acetate, formic acid and TFA.
Vlad
Posted: Mon Sep 12, 2005 8:47 pm
by zydus
dear friends,
thankx everybody again,
the retention time of levidopa and carbidopa are 2.5min and 3.9. the initial buffer acn ratio is 98:2 and i am using 250 mm column. will it be helpful to start with 100%water?
with regard
dhiren :lol:
Posted: Mon Sep 12, 2005 8:56 pm
by SIELC_Tech
Void for 250 mm column at 1 ml/min is approx. 2.5 minutes. One of your peaks is in void/ not retained. 100% vs. 98% is not going to change a lot. Any matrix effect will interfere with your first peak (2.5 minutes). May be AQ type of column will help a little.
Posted: Mon Sep 19, 2005 4:37 am
by ym3142
I am not familar with your analytes but as general, you can try the following to help holding:
1) some columns are designed for very hydrophilic compounds;
2) use methanol instead of ACN;
3) use ion pairing reagents with long alkyl chain, high concentration, and a right pH if necessary.
good luck,