lmh » Fri Sep 12, 2014 4:16 pm
													
 
						
					 
					
						I would query why you want a triple quad if the main goal is identification of compounds in unknown samples. By unknown samples, I assume you'll want to identify compounds where you cannot predict in advance what they might be, even to the extent of having a long list of likely possibilities.
Triple quads are good where you have a list of possibilities. They are best when set up to look for a suite of previously-defined mass transitions, each corresponding to some known compound. They are often very poor at collecting full-scan data, and of course they don't offer accurate mass, which is one of the most useful tools for identification of unknown compounds. They can collect MS2, for searching against libraries (if you have one...), but since they have to choose what parent to select for fragmentation, and they're then scanning the fragment ions (the process they're not so good at), you won't get spectra for every ion in the run (it's a bit random what you do get), and the spectra won't be of particularly good quality. (On the other hand, being CID in a collision cell, the spectra may be more rich than you'd get from many ion-traps using resonant excitation, where you should only see single fragmentation reactions - no fragments of fragments; ion trap MS2 spectra are often dismally dull for a library search).
Triples aren't cheap, so your budget can't be too small. Have you got enough cash to consider a ToF instrument for the accurate mass (particularly, it's worth asking the suppliers whether your budget would stretch to an ex-demo instrument if a new one is out of range)? You could even consider an Orbitrap of some flavour.
Whatever happens, get the manufacturers to demonstrate, and provide you with data from test-samples you supply. It's always reassuring to see that the instrument you propose to buy can generate the sort of data you'd like to see...