Strange area pattern in MHE-GC

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi all
I am setting up a MHE method with 4 injections on an Agilent HS sampler (model 4557) for a residual solvent method (ethanol, toluene and acetone in API).
My areas for the ethanol peak does not behave as expected: The first peak is low, the second and third peak are higher than the first peak and have same size, and the fourth peak is lower than the first. The method I use has previously been working well on a Perkin Elmer HS sampler.

Please suggest way to improve the method on the Agilent HS sampler.

Cheers,
M
It sounds like your autosampler is not going through the pressurize-depressurize cycle it needs to go through for MHE.
Hello

What you have is poor reproducibility. Possible reasons:
-leak in HS sampler (sample loop, pressurization valve, vent valve).
-leak in GC (transfer line-inlet tubing, septa, vent trap/tubing, valves or EPC module)
What you need to do is to find a leak.

Regards

Tomasz Kubowicz
3 posts Page 1 of 1

Who is online

In total there are 34 users online :: 1 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry