HPLC column theoretical plate count
Posted: Wed Sep 03, 2014 2:05 am
by loannguyen429
I'm doing method validation using Discovery BIO Wide pore C18 column. The column was used several times 4 years ago. I followed the SOP exactly, the peak shape and resolution was very good, but the system suit failed on the theoretical plate count. The plate count was low. I'm wondering if the problem is due to the column itself. Anything else you suggest I should do now?
Re: HPLC column theoretical plate count
Posted: Wed Sep 03, 2014 2:56 am
by tom jupille
I think we may have a terminology problem here. Are you actually validating a method or running an existing validated method?
First of all, I don't think you can "fail" a system suitability test during validation, because defining system suitability criteria is part of the validation process.
Second, what do you know about the history of that column? Has it been used for other things? Has it been properly stored? You cannot properly validate a method using a column of dubious provenance; there is no guarantee that previous use has not changed its properties.
Third, an important part of validation is determining the robustness of the method, including robustness with respect to column chemistry. That means that you should be running with at least two (preferably three) columns from at least two identifiably different batches or lots.
If you are simply trying to verify or run the method (rather than fully validating it), yes it is possible that the column has been damaged either through abuse or improper storage. That said, it is possible that the problem is something physical such as extra-column volume from large-bore tubing or an improperly seated tube end in a connecting fitting. If this were my problem, I would check all the connections and then re-test with the manufacturer's test procedure on the CoA that came with the column. If it fails, get a new column. Columns are cheap, your time is valuable! It's very easy to spend more time than the column is worth.
Posted: Wed Sep 03, 2014 6:33 am
by Gerhard Kratz
In addition to Tom I will make it short: Purchase a new column!!!
We can only speculate what happened in the past 4 years with the packing material during storage.
Posted: Wed Sep 03, 2014 7:21 pm
by loannguyen429
I'm actually run the existing method for API. Thank you for your advice!