Flow cell 2998 Analytical

Posted: Tue Sep 02, 2014 10:49 pm
by ashishsaksule
If anyone using HPLC Waters e 2695 system and having trouble with UV detector flow cell.
We are facing problem in e 2998 flow cell, tubing seems to be blocked.
Any suggestions?
Anyone have the same used part and want to trade?

Re: Flow cell 2998 Analytical

Posted: Wed Sep 03, 2014 6:36 am
by Gerhard Kratz
Give a call to the Waters Service guys. Disconnect the tubing from the detector. Connect a syringe to the inlet and try carefully to get some Isopropanol through the cell. Good luck