Advertisement

Flow cell 2998 Analytical

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
If anyone using HPLC Waters e 2695 system and having trouble with UV detector flow cell.
We are facing problem in e 2998 flow cell, tubing seems to be blocked.
Any suggestions?
Anyone have the same used part and want to trade?
Give a call to the Waters Service guys. Disconnect the tubing from the detector. Connect a syringe to the inlet and try carefully to get some Isopropanol through the cell. Good luck
Gerhard Kratz, Kratz_Gerhard@web.de
2 posts Page 1 of 1

Who is online

In total there are 20 users online :: 1 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry