Chemstation A.06.04 - HP 1050 problem

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

5 posts Page 1 of 1
I have on old HP 1050 LC with a Chemstation A.06.04. My problem is that I can see the chromatogram on the monitor (using a VWD), but the signals of the VWD isn't saved to the data file, e.g. the file vwd1a.ch isn't created in the data directory. The VWD channel is of course assigned to the method.
Could you help me with this problem?
Hello,

on a DAD this is the case when you don't mark a box in the DAD Signal settings for "Save".

Then the Data-File isn't created.

This should be the same on a VWD.


Edit: Hm, I see, that on a VWD there is no box. :(
I've searched through the method settings and practically all the instrument settings in Chemstation and found nothing which can be in connection with the problem.
Do you have all the remote cables hooked up and not just the GPIB?
I'll check it (on Monday). They should be attached. By the way, commands send out to VWD from the Chemstation command prompt are going through remote cables or GPIB? 'cos sending e.g. the "lampall on" command is working and the lamp turns on.
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry