Low FID signal
Discussions about GC and other "gas phase" separation techniques.
8 posts
Page 1 of 1
I have an Agilent 6890 with dual injection ports and dual FIDs. When I inject from the front injection port to the front FID, I get a solvent peak of about 3500 pA. When I inject from the front injection port to the back FID I get about 300,000 pA for the solvent peak (i.e. much higher). I swapped the jet and collector assembly from the front FID to the back FID and I get 300,000 pA. From this, I gather the injection port is ok and the jet/collector assembly is ok. Any idea on what could be wrong? Something with the electronics? Measurements are being done through ChemStation through an ethernet connection to the GC. Thanks!
-
- Posts: 82
- Joined: Tue Nov 27, 2012 11:26 am
After you swapped the jets and collector and got 300,000 pA, was that injecting from the front inlet to the front detector (which now has the parts that were on the back detector? If so, that says there is a problem with the original front detector parts. It's possible your original front detector jet is slightly clogged. I've seen this before but not with such a large loss of signal. How are your retention times? If the retention time is longer for the injection in which you got 3500 pA, this also suggests a clogged jet.
-
- Posts: 10
- Joined: Wed Nov 28, 2012 12:25 am
Argh, nevermind. The air flow rate to the front detector wasn't set right (way too low) and I didn't notice it. We never change those settings, so it didn't come to mind to check it. All good now.
-
- Posts: 45
- Joined: Sun Jun 23, 2013 9:34 pm
Which begs the question: what is the optimal air flow and H2 flow on a FID
or does it really matter as long as the Flame is on and set at the correct Temperature?
or does it really matter as long as the Flame is on and set at the correct Temperature?
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
30/300
35/350
40/400
45/450
The best way is to check analyte(s) response vs. different H2/Air ratio and pick the best one
Regards
Tomasz Kubowicz
Normally Hydrogen/Air - 10-12% which in practise gives typical values:Which begs the question: what is the optimal air flow and H2 flow on a FID
or does it really matter as long as the Flame is on and set at the correct Temperature?
30/300
35/350
40/400
45/450
The best way is to check analyte(s) response vs. different H2/Air ratio and pick the best one
Regards
Tomasz Kubowicz
-
- Posts: 10
- Joined: Wed Nov 28, 2012 12:25 am
We have ours set to 40, 400 and 45 mL/min for H2, air and N2 (makeup).
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hi
If you're using gas generators it doesn't matter but if you have gas cylinders you can calculate H2 and Air consuption.
Regards
Tomasz Kubowicz
It is all about method development...We have ours set to 40, 400 and 45 mL/min for H2, air and N2 (makeup).
If you're using gas generators it doesn't matter but if you have gas cylinders you can calculate H2 and Air consuption.
Regards
Tomasz Kubowicz
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
You can rig up one of these:
THOMASON, M.M., BERTSCH, W., APPS, P.J. and PRETORIUS, V. 1982. A simple device for the optimisation of flame ionisation detectors and determination of absolute sensitivity. Journal of High Resolution Chromatography and Chromatotgraphy Communications 5: 690-692.
But you will see only marginal improvements over what the manufacturers recommend for their instruments.
Peter
THOMASON, M.M., BERTSCH, W., APPS, P.J. and PRETORIUS, V. 1982. A simple device for the optimisation of flame ionisation detectors and determination of absolute sensitivity. Journal of High Resolution Chromatography and Chromatotgraphy Communications 5: 690-692.
But you will see only marginal improvements over what the manufacturers recommend for their instruments.
Peter
Peter Apps
8 posts
Page 1 of 1
Who is online
In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 18 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 18 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science