Does ammonium formate offer a different selectivity as compared to ammonium acetate in HILIC mode?
Thanks.
mwahab@ualberta.ca
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Any complexing acid, like citric acid, is going to be a troublesome additive. It will most likely give huge system peaks. Any additive which interacts too strongly with a stationary phase is perhaps going to be problematic.On a different topic - but still in the realm of "HILIC Additives" - does anyone know if Citric Acid, or its salts, can be used in HILIC (i.e. are they soluble in the high organic). I've tried to look this up, but I don't see this one listed on either the "do use" or the "don't use" list (at least in the references I have on hand).
Thank You for any feedback.
This is interesting. What is your Obelisc N column chemistry (in generic terms if you wish). Thanks.The effect of ammonium formate or ammonium acetate is more pronounced in our mixed-mode columns, since we are exploring HILIC and ion-exchange mechanisms. pH of you buffer not only change ionization state/polarity of your analyte but also changes polarity and ion-exchange properties of the stationary phase (Obelisc N). You can even achieve change in order of elution when you go from ammonium formate pH 3 to ammonium acetate pH 5. You can use this tool to adjust selectivity of your separation.
Thanks. Just curious, if the positive charge is so close to the silica surface (as shown), how is this bonded phase stable with respect to hydrolysis or is it a cross linked polymer coating?Here is surface chemistry:
http://www.sielc.com/Products_Obelisc.html
That is interesting.The higher organic solubility. That said, experience has shown that having a lot of phosphate in the mobile phase lets you do some things that you can't do as effectively with other anions. For example, in ERLIC, starting the chromatography with sodium methylphosphonate in the mobile phase results in heightened selectivity for retention of phosphopeptides. Subsequently running a gradient to 0.2 M TEAP elutes them all, even ones with 4 phosphates. If you tried that with a volatile salt such as ammonium formate, you'd have trouble eluting all of the peptides with 3 phosphates, to say nothing of the ones with 4.
Historical note: When I started using TEAP, it was because it was transparent at 220 nm. At that time I only had access to an absorbance detector.