OK- quick re-do, this time metering for resistance instead of diode check. Dummy me.
With the switch set to "Filament 1," I get 0.5-0.6 ohms across the solder connections for filament 1, and 1001 ohms for filament 2.
Switched to "Filament 2," I get about 0.9 ohms for filament 1, and 1001 ohms for filament 2.
The difference in filament 1 is probably due to the inability to get an ideal contact on those old solder joints.
So- in this instance, would either filament be bad?
As for why I'm doing all this- I am getting some wonky readings, trying to get this old MS back in service.
When I autotune (long or short), I get "Cannot achieve constant peak widths" to "Fault Status 2" to "gain and offset = 0 and peaks are too narrow".
With manual tune, I get 
With autotune, I get "Tune error- there is not enough signal to begin the tune".
I get the same with both front and back filaments.
There IS PFTBA in the vial.
I was advised to dip the coils, and got the following:
RFPA = 10, voltage 0.011 VDC
RFPA = 100, voltage 0.044 VDC
RFPA = 300, voltage 0.118 VDC
RFPA = 600, voltage 0.240 VDC
I did the above at my mentor's recommendation. My suspicion is that the electron multiplier is dead, but he didn't think this was necessarily the case. However, my mentor has since disappeared, and I've been trying to snoop out what's wrong ever since.