Single largest Impurity
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
8 posts
Page 1 of 1
I have six replicate samples. My first three trial has largest impurity at different retention time compared to my last three samples. Let say on my first three trial, I'm getting it at 4 minutes, whereas on my last three samples, I'm getting it at 10 minutes though it has 4 minutes impurity also but lesser compared than my 10 minutes. Am I right of saying, that what I'm just going to consider on my calculation are those largest impurity coming regardless of what retention time they have? thanks.
-
- Posts: 775
- Joined: Mon Dec 01, 2008 9:59 pm
make sure that what ever you see is not impurity coming from a previous run.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
If you inject the same sample 6 times retention time for impurity should be the same (even if peak area is not reproducible). It looks like you have problem with pump (flow or eluents mixing). I'd suspect: ball valves, cartidges in inlet valve or gradient valve.
Regards
Tomasz Kubowicz
If you inject the same sample 6 times retention time for impurity should be the same (even if peak area is not reproducible). It looks like you have problem with pump (flow or eluents mixing). I'd suspect: ball valves, cartidges in inlet valve or gradient valve.
Regards
Tomasz Kubowicz
-
- Posts: 95
- Joined: Fri Nov 30, 2012 5:18 pm
what i mean here is six sample preparations? the first three sample prep has largest impurity on 4 minutes let say but the remaining three is at 10 minutesHello
If you inject the same sample 6 times retention time for impurity should be the same (even if peak area is not reproducible). It looks like you have problem with pump (flow or eluents mixing). I'd suspect: ball valves, cartidges in inlet valve or gradient valve.
Regards
Tomasz Kubowicz
-
- Posts: 512
- Joined: Tue Jul 19, 2011 9:26 pm
Is it six preparations of the same sample or six preparations of six different samples??what i mean here is six sample preparations? the first three sample prep has largest impurity on 4 minutes let say but the remaining three is at 10 minutes
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
You have 2 possible situations:
1.You have six different samples. For 3 injections you have ret. time for IMP about 4 min and for other 3 injections ret. time is about 10min. It means that you have 2 different impurities. Normally under the same conditions retention time is consistence for the same analyte (if pump is working fine!)
2.You have six injection of the same sample (prepared separately). Retention time for impurity should be the same even if concentration (peak area) is slightly different because of errors in preparation step (extraction, cleanup etc.). If retention time drifts you have problem with pump (flow, mixing, gradient)
Regards
Tomasz Kubowicz
You have 2 possible situations:
1.You have six different samples. For 3 injections you have ret. time for IMP about 4 min and for other 3 injections ret. time is about 10min. It means that you have 2 different impurities. Normally under the same conditions retention time is consistence for the same analyte (if pump is working fine!)
2.You have six injection of the same sample (prepared separately). Retention time for impurity should be the same even if concentration (peak area) is slightly different because of errors in preparation step (extraction, cleanup etc.). If retention time drifts you have problem with pump (flow, mixing, gradient)
Regards
Tomasz Kubowicz
-
- Posts: 95
- Joined: Fri Nov 30, 2012 5:18 pm
so let say i fall over situation number 1 for my first three sample prep, i also have 10 minutes but 4minute impurity has larger area. so which one should i report as single largest? thanksHello
You have 2 possible situations:
1.You have six different samples. For 3 injections you have ret. time for IMP about 4 min and for other 3 injections ret. time is about 10min. It means that you have 2 different impurities. Normally under the same conditions retention time is consistence for the same analyte (if pump is working fine!)
2.You have six injection of the same sample (prepared separately). Retention time for impurity should be the same even if concentration (peak area) is slightly different because of errors in preparation step (extraction, cleanup etc.). If retention time drifts you have problem with pump (flow, mixing, gradient)
Regards
Tomasz Kubowicz

-
- Posts: 512
- Joined: Tue Jul 19, 2011 9:26 pm
I'm not really sure if I got it
You have six independent samples, right? It's NOT one sample with six independent preparations?
Your API(s) peak(s) show consistent retention times over the whole sample set?
So you have six different samples of which some show different impurity profiles...and what's the problem, now? If your SOP says to report the largest impurity, then you should report the largest impurity. If the largest impurity is a different one in different samples, then, well, you have to report different impurities for the different samples.

You have six independent samples, right? It's NOT one sample with six independent preparations?
Your API(s) peak(s) show consistent retention times over the whole sample set?
So you have six different samples of which some show different impurity profiles...and what's the problem, now? If your SOP says to report the largest impurity, then you should report the largest impurity. If the largest impurity is a different one in different samples, then, well, you have to report different impurities for the different samples.
8 posts
Page 1 of 1
Who is online
In total there are 4 users online :: 3 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Bing [Bot], Google [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Bing [Bot], Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science