Extractable from rubber stopper
Posted: Fri Aug 22, 2014 11:18 am
by Mattias
I struggling to identify an extractable from a rubber stopper by LC/MS (ESI, QTOF, positive mode, with lockspray)
The unknown has mass of 269.104 [M+H] and fragments into 135.06, 107.06, 92.05 and 80.04.
It has UVmax at 228 and 277 nm (aromatic)
Has anyone seen this one before?
Re: Extractable from rubber stopper
Posted: Sat Aug 23, 2014 3:26 am
by JMB
Mattias,
What is the accurate mass of m/z 269, and from that the possible elemental compositions ?
Possibly m/z 135 is [CH3OC6H4C=O]+
and m/z 107 [CH3OC6H4]+
JMB
Posted: Sat Aug 23, 2014 10:31 pm
by dhdh
This probably isn't it, but I notice there are some masses in common (or off by 1) with the the rubber additive benzothiazole. Prominent molecular ion 135 on GC-EI-MS, and fragments 108, 91, 82 and 69.
Posted: Mon Aug 25, 2014 3:37 am
by JMB
This raises the issue of whether m/z 269 is indeed [M+H]+ or is [2M+H]+ ?????
JMB
Posted: Mon Aug 25, 2014 3:39 am
by JMB
Also, 2-mercaptobenzothiazole is used as a vulcanization agent for rubber (Wikipedia).
Can you detect presence of S in m/z 269 ion, i.e. enhanced signal intensity at m/z 271 ??
JMB
Posted: Mon Aug 25, 2014 9:45 am
by Mattias
JMB
That is a good question! Even without MS/MS the 135.05 peak is much larger than the 269.10 peak. I also see some 291.1 peak which then would be [2M+Na].
But [M+H] of benzothiazole should weigh 136.0 and [2M+H] should weigh 271? Could it be possible to knock of an electron from the aromatic ring and get the [M+] ion instead?
Posted: Mon Aug 25, 2014 12:28 pm
by Mattias
Also, when I perform fragementation on m/z 269, I do not see any fragments between 269 and 135.
So my molecular ion is most probably 135 (and not 269).
Posted: Mon Aug 25, 2014 12:41 pm
by Mattias
The closest hit on
135.0552 (determined with accurate mass) is
C7H7N2O (protonated), calculated mass
135.0558.
Google on C7H6N2O + rubber chemical = 2-hydroxybenzoimidazole. I cannot find any ms/ms data on this one, but I have just ordered it
Posted: Mon Aug 25, 2014 3:42 pm
by JMB
Mattias,
2-OH-benzimidazole is a good possible, although I cannot get m/z 80 from it.
Do you have these two sources; Massbank is very nice.
http://www.massbank.jp/index.html
http://www.lc-ms.nl/contaminants.htm
JMB
Posted: Tue Aug 26, 2014 7:20 am
by Mattias
http://www.massbank.jp/jsp/Dispatcher.j ... 116&site=0
This is the fragmentation of 2-amino benzimidazole. It weighs one mass unit too little, but it fragments into 65, 80, 92 and 107. Exactly the fragments that I observe.
Will be very exciting when I get the reference substance
Posted: Tue Aug 26, 2014 9:03 am
by Mattias
http://rd.springer.com/article/10.1007% ... 11-1104-0#
In the bottom of page 305 the fragmentation of 2-OH-benzimidazole in ESI-MS is described. It fits perfectly to my observations.
But I assume that the fragmentation of small molecules can be quite similar, final confirmation will have to be done with retention time, UV-spectrum and MSMS fragmentation.
It is fun to work
Posted: Tue Aug 26, 2014 2:54 pm
by JMB
I retired a couple of years ago, but still try to keep my brain ticking over by looking at these MS problems.
So much for..............
"Possibly m/z 135 is [CH3OC6H4C=O]+
and m/z 107 [CH3OC6H4]+"
Enjoy your work !!
JMB
Posted: Thu Aug 28, 2014 12:15 pm
by Mattias
I have now confirmed that the peak is 2-hydroxybenzimidazole
It was a great help that you (JMB) made me aware that the 269.1 peak was not the molecular ion! Otherwise I would probably be still stuck in this problem.
Posted: Thu Aug 28, 2014 3:45 pm
by JMB
Mattias,
Excellent, that was your starter question for 10 points..............
JMB