Page 1 of 1

Please help: Weird peak of separation of 5 amphetamine drugs

Posted: Wed Aug 20, 2014 1:14 pm
by fuh123
I'm using a previous established LC-MS/MS method for detection of 5 amphetamine drugs. AB Sciex QTrap 5500 with Shimazdu LC.
Mobile Phase A 0.1% Fomic Acid in H2O Mobile Phase B ACN.

I got weird peak as shown in the pic. The 1st peak (red one) is amphetamine, which was really broad. I was wondering if something wrong with the column? I don't think MS part wrong. So might be the LC part problem. Do I need to flush the column more often?

https://drive.google.com/file/d/0B7AR3a ... sp=sharing

Image

New to LC-MS world. Please make your comments. Thanks!

Re: Please help: Weird peak of separation of 5 amphetamine d

Posted: Wed Aug 20, 2014 2:53 pm
by tom jupille
My first guess would be a gross overload. Easy to check:dilute 1/10 and reinject.

Re: Please help: Weird peak of separation of 5 amphetamine d

Posted: Thu Aug 21, 2014 10:22 am
by carlo.annaratone
I would say overload as well

Re: Please help: Weird peak of separation of 5 amphetamine d

Posted: Sun Sep 07, 2014 1:54 pm
by fuh123
Thanks for the reply!

Updates: It was not an overloading issue. Turned out the column related. By switching to a new column, peaks were perfect. And ACN was not a good mobile phase. MeOH is better for amphetamine type compounds. Just FYI.

Re: Please help: Weird peak of separation of 5 amphetamine d

Posted: Sun Sep 07, 2014 2:43 pm
by tom jupille
Thanks for the update!