Problem with generic Kromasil C18

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi,
we are using Kromasil C18 column provided by a french society who pack the phase from EKA, it's the third column we buy and each time we get a pressure increasement at about 1000-1200 injection ex: 165b @0.7ml/min, 30-70 ACN-water. The column is fitted with online filer and c18 precolumn.
I wonder if the problem is the packing method or not, may i change my provider ? :?
Thanks

If you have a column with pressure too high to use anymore (nothing left to lose) try changing the inlet frit. If that solves the pressure problem, then the column vendor is not at fault. To prevent the problem, try a smaller-porosity pre-column filter; try more thorough sample filtration; try changing your pre-column more frequently.

If changing the inlet frit does not reduce the pressure, try changing the outlet frit. If that solves the problem, it may be a vendor-related problem.
Mark Tracy
Senior Chemist
Dionex Corp.
2 posts Page 1 of 1

Who is online

In total there are 19 users online :: 1 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry