Losing connectivity on shutdown

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
Hi all,

We are runing new agilent 1260 systems on Empower 3 with SR1. We have the correct drivers installed and were able to successfully set the systems up and view the control panel on run samples. However if we were to reboot the client they are connected to we lose the control panel and the only way to fix it is to delete the systems and reinstate them from the start.

We are still able to ping the IP's on the client and in the node properties when we scan instruments it says yes but doesnt contain the information in the details column.

What i think it may be related to is the fact that in configure dhcp we get no addresses up so we add them manually with the mac address. Any light anyone could shed on this i would really appreciate, Thank you
The client should have a static IP address on the domain of empower not DHCP which reassigns each time.
2 posts Page 1 of 1

Who is online

In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry