Air Leak in GC-MS
Posted: Wed Aug 13, 2014 11:57 am
by rsa01
Hi all,
My GC-MS system having minute air leak now i want to check where exactly that leak from. Please give some Suggestions.
Regards
Ashok
Re: Air Leak in GC-MS
Posted: Wed Aug 13, 2014 12:24 pm
by kubowicz.tomasz
Hello
Short spray at the suspect points (inlet, column connection, transfer line) could do the trick. You can use Argon (ion 40) or tetrafluoroethane (69, 83)
Regards
Tomasz Kubowicz
Posted: Wed Aug 13, 2014 1:47 pm
by rsa01
Hi Tomasz Kubowicz,
Any special spray is available in the market to find out leaks?please advice.
Regards
Ashok
Posted: Wed Aug 13, 2014 2:10 pm
by Bigbear
Try canned aiir, there is a label saying what's in it, scan for it's major ion when spraying.
If the air/water is <20% Agilent says you're good to go.
Posted: Wed Aug 13, 2014 2:14 pm
by kubowicz.tomasz
Hello
I would not recommend air. If your leak is massive you can burn filament in a minute.
Regards
Tomasz Kubowicz
Posted: Wed Aug 13, 2014 2:34 pm
by Peter Apps
Hello
Regards
Tomasz Kubowicz
"Canned air" is not air, it is one or other freon. Even if it was air it could not be worse for the filament than the atmosphere that is all around. And the OP already said that the leak is minute.
Posted: Thu Aug 14, 2014 6:52 am
by BMU_VMW
Might be a silly question, but have you already changed or re-installed your septum?
Posted: Sat Aug 16, 2014 6:05 am
by rsa01
Hi Mr.BMU_VMW,
I already changed septum,liner,liner O ring. actually as per agilent they recommended air level <10% & water level <20%.But my instrument showing air level nearly 11%.I think this may be due to some minute leak. I am having small doubt this may affect the analysis result. Thats why i want to arrest the leak.
Regards
Ashok