Varian GC EFC type 4 pressure problem

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Please help Varian GC with EFC type 4,Helium with packed column,The column pressure set point is 21.7 psi and flow of 60 ml/min,please can some one help me troubleshooting this issue of GC showing not ready due to EFC pressure of 21.9 psi (Set point 21.7 psi) how can i over come this and is there a way to adjust.
For your information room tempearture increased slightly due to AC problem on this particular date.
Thanks.
Maybe the inlet of your packed column is blocked. Disconnect the column and check flow and pressure.
Gerhard Kratz, Kratz_Gerhard@web.de