Recommended Criteria for Accuracy
Posted: Wed Aug 13, 2014 7:08 am
by jayroseville58
As per ICH, Accuracy is being demonstrated by performing 3 replicates each of three separate sample concentrations.
Question:
Let say, I have an Assay Specification of 80% -120%, What would be the recommended criteria for its percent recovery (on Accuracy test). Does ICH has clear guideline on this?
Can I set it as 95% - 105% percent recovery?
Thanks.
Re: Recommended Criteria for Accuracy
Posted: Thu Aug 14, 2014 10:51 am
by gtma
That seems reasonable to me given your spec is relatively wide. The tightest I would set the accuracy limit is 97.0% to 103.0% for such a wide specification. Please made sure you take into consideration the potency of the drug and other variabilities e.g. other analytical method variabilities, potency loss on stability, and manufacturing variability.
Posted: Thu Aug 14, 2014 8:27 pm
by Blazer
ICH does not have a recommendation for accuracy recovery. I would let your prevalidation data drive your acceptance criterion for your protocol.
Posted: Fri Aug 15, 2014 11:15 am
by lmh
The two values interact. Logically, your product specification needs to be aware of your accuracy of measurement. The thinking should be like this:
(1) For the intended application of this product, a variation more than (for example) +/- 20% would be bad.
(2) Include a general safety-factor if desired. Maybe if 20% deviation has seriously bad consequences, we would prefer to reject product that deviates by more than 10%.
(3) If we think the analytical precision and accuracy allow a 5% error, then if we specify +/- 10%, a measurement of 10% could actually be as bad as 15%, so instead we should insist on a smaller analytical error-window such that analysis still guarantees that the product is within the safety-window we specified. Therefore if we want to guarantee +/- 10%, we can specify our product as +/- 10% if our analysis is perfect, or perhaps +/- 5% if it's not...
What accuracy and precision you get is going to depend on the instrument, the analyte, the level at which you're analysing etc. etc.; it has to be good enough for purpose, but for once I agree with ICH in not specifying a rigid number.