%RSD of Standard Preparation

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
A value of NMT 2.0% of %RSD for five injections of standard preparation is indicated as per USP. We usually do it by injecting one vial six times. Is it really necessary that we are going to consider the five consecutive injections? Let say, we have six injections (1 to 5 or 2 to 6) in order to test the suitability of the HPLC system? Thanks.
Hello

In my opinion it is better to inject standard from different vials one by one instead of injections from the same vial. Possible problems you can avoid:
- micro-dilution - small amount of solvent outside the needle can dilute standard in the vial
- if your solvent is volatile and injection volume is quite big , you can have headspace phase in the vial and concentration of your standard can be higher after couple of injections


Regards

Tomasz Kubowicz
Hello

In my opinion it is better to inject standard from different vials one by one instead of injections from the same vial. Possible problems you can avoid:
- micro-dilution - small amount of solvent outside the needle can dilute standard in the vial
- if your solvent is volatile and injection volume is quite big , you can have headspace phase in the vial and concentration of your standard can be higher after couple of injections


Regards

Tomasz Kubowicz
Okay so let say we are doing what you suggested and I have vial number 1 to 6 and all of a sudden my vial number 5 has a problem. Should I not consider vial 1 to 4 plus vial number 6 for the calculation of %RSD for my System Suitability Test? USP mentions five replicates at NMT 2.0%. Does it mean consecutive vial replicates? Thanks.
Hello

Try to imagine situation...
You do what you mentioned (you skip vial no5 and you take 1-4 plus 6). You calculate RSD and then you run your sample.
What if you'll have the same problem as for vial no5 (wrong volume injected) ? You won't be able to tell if your sample was injected properly...You'll calculate sample concentration and it'll be wrong.
So if one of your consecutive injection is causing problem for RSD calculations, I would say that there is something wrong with your LC system (sampler).
I assume of course that peak integration is out of the question...

Regards

Tomasz Kubowicz
We inject five from the same vial before samples are injected, plus another injection at the end. Both initial SST and on-going SST are calculated, both must pass.
Thank You all for the responses
6 posts Page 1 of 1

Who is online

In total there are 38 users online :: 2 registered, 0 hidden and 36 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 36 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry