Standard and Sample Concentration on my Assay
Posted: Tue Aug 12, 2014 5:18 am
by jayroseville58
What is the acceptable difference between my standard and sample concentration on my Assay Test. Let Say, my sample concentration is 0.12mg/ml and my standard is 0.15mg/ml. Is it still okay?
Re: Standard and Sample Concentration on my Assay
Posted: Tue Aug 12, 2014 7:48 am
by Gerhard Kratz
In my opinion it should be the same concentration I sample and standard.
Otherwise you need to do a calibration curve every time.
Posted: Tue Aug 12, 2014 9:30 am
by kubowicz.tomasz
Hello
In typical QC assay run standard should have "the same" (in reality - very close to) concentration as your sample. It is simplification but in fact it is calibration curve with two points (0 and Conc. of std)
So it depends on your limits for assay...very common limit is 98.5-101.5%
Regards
Tomasz Kubowicz