I have never had a GC/MS system that I didn't see some air and water and other gasses present. Normally I get 44 CO2 and 40 Ar along with 32 O2, 28 N2 and 18 H20.
http://www.sisweb.com/referenc/source/exactmas.htm looks like 19 could be F or possibly H2O(17) but that would be unlikely since O(17) has such low abundance.
The way to tell if you have a leak big enough to worry about is to look at the abundance of those ions versus m/z 69 from the PFTBA. If they are greater the 5% you could have a leak, greater than 10% and you will begin to see problems from oxidization in the source over time. If they are down around 1% they you have a seal on the system that is nearly perfect, at least as far as I have ever experienced.