convert capillary 1100 hplc to higher flow for porous shell

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi
I have a capillary 1100 hplc pump which i'd like to work at higher flow rates- say 200-500ul/min using porosshell or halo etc type columns. I wonder what you would suggest about doing this conversion- will i have to replace the microrotor seal and microrotor with a conventional one, replace all peeksil fittings with metal tubing,work in normal flow modeinstead of capillary, utilise a column oven at elevated temps to reduce vbackpressure?
I'dreally appreciate your ideas- there are examples for conventional flow 1100 conversions but none for capillary flow 1100 systems
Just change the tubing with larger ID of PEEK capillaries. Maybe you can use still your injection system and the flow cell of your detector.
What is the flow range of your pump system?
Good luck
Gerhard Kratz, Kratz_Gerhard@web.de
2 posts Page 1 of 1

Who is online

In total there are 602 users online :: 1 registered, 0 hidden and 601 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 601 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry