Waters 717 setup - please help!
Posted: Tue Jul 29, 2014 1:55 pm
By searching, I noticed a lot of members here have Waters systems successfully set up.
We just got in, as a donation from another lab in our area, a supposedly-working system.
It was a pain to get the software online, seeing as they wiped the hard drive (and you know those systems are single use), but the hardware should have been far easier.
Holy mother, it was far worse. The instrument was set up, and the lab sent someone to go observe and get trained, but unfortunately, he didn't make note of how the cables were connected. We have some grainy pictures of their setup, but I'm really having trouble playing around with it.
What we have:
Waters 717 plus Autosampler
Waters 515 HPLC Pumps
Waters Temperature Control Module
Bio-Rad Electrochemical Detector Model 1640 (x3)
Guard Column and thermal case
Waters BusSAT/IN Module
Waters Pump Control Module
We have the special card with the female-RS232 and IEEE ports. When I spoke to the Waters representative, she said that the IEEE was all we needed. I'm told that the system, because of the Bio-Rad detctor, needed the special wires to be plugged in to the back of the autosampler.
But I don't know how each of the cables should be plugged in. If anyone has good photos from their lab of their setups, that would be appreciated. There is another group that has a Waters detector, but it may or may not be fully functional (that, and it is connected to their machine, actually under their controller, and I don't want to disconnect anything)
What I have (recalling from memory):
Computer has IEEE to the autosampler.
The 515 is connected to the Pump Control Module by a square cable. The Pump Control is then plugged into the back of the autosampler by the IEEE.
What we have:
