I had two columns sitting in the compartment one a Phenomenex Rezex Monosaccharide Ca2+ (ion exchange) and the other a Zorbax XDB-C8.
Both were used for sugar analysis, the eleunt for the Rezex was 18mohm H2O and the C8 was 75% ACN, they were probably sitting in this mobile phase for 2 years. And there is a possibility the columns were sitting in the wrong type of mobile phase for that period. I spent ages flushing out with IPA and then the mobile phase. Initially the RID kept getting unbalanced even after purging the reference cell multiple times. Seems ok now but I'm still getting funny readings, I'd like to show some chromatograms but don't know if it's possible to upload photos here.
Anyway the Rezex seems to have a non-fluctuating baseline though it slowly changes but it's a straight line. However I put in sugar standards 1% of fructose, glucose and sucrose and the peaks overlap a lot which shouldn't occur on a monosaccharide column, also a 1% maltose standard produced a badly tailed peak. The XDB-C8 has a very wavy baseline, up and down which is noticeable at the 100,000 nRIU scale. I switched the XDB-C8 to a Zorbax C18 and same thing with the wavy baseline as the XDB-C8.