I can see where you select the detector minus column comp. However, I am having trouble finding where you actually perform the column compensation run.
Brad
Advertisement
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
Thanks for the response. I guess I should have noted this is a Agilent 6850 connected to a computer running EZchrom Elite 3.3.1. Is it possible to do the column compensation run in EZchrom Elite?Hello
If you're using Agilent GC do as below:
You need to run it from GC keybord.
Regards
Tomasz Kubowicz
Hello
I'm not sure if you can run it in EZCh but check detector signal options...you should have dropdown menu with
-signal
-signal - column compensation
-column compensation
If not, use hand controler from GC 6850 (instant pilot) to run it
Regards
Tomasz Kubowicz