CTC/Gerstel Autosampler alignment problem
Posted: Mon Jul 28, 2014 9:28 am
I'm responsible for the upkeep of the Agilent 6890/5973 GC-MS in our lab. The actual instrument is working perfectly, however there is an issue which keeps creeping up with the autosampler - which is a Gerstel MPS autosampler (made by CTC I think? It's tricky to find information on these as I think they could be rebranded?!).
When running samples, every so often the syringe does not align with the septa on top of the vials and as a result, the needle is pushed into the side of the cap, resulting in a bent needle. We recently had to order a new pack of syringes as we were getting through one every 2 weeks and have moved back to manual injections until the problem is solved.
I've been through the calibration process with the control panel however I am only able to calibrate to each corner position on the vial rack, not individual positions. It has been aligned perfectly with these positions so I am mystified as to why vials in the intermediate positions tend to result in this off alignment.
Having recently inherited responsibility for the machine (I am a synth. chemist primarily but keep things running with the GC!) I believe that this was an issue in the past however nobody is able to tell me how the issue was fixed - indeed when I started back in October we were using manual injections and it is only recently that I've managed to get the robot up and running again. Our Mass spec. department have had a look however cannot offer any suggestions.
To summarise, is there a way to access calibration points for individual positions? It will be time consuming but it's something I'm prepared to do. Other issue could be that the vial tray is warped however as far as I can see with a spirit level this is not the case.
Many thanks for your help in advance,
Geecee
Geecee