Posted: Wed Jul 23, 2014 5:47 pm
by chloesabrina
I have started in a new laboratory and they had a lot of preventative maintenance not being performed so I have been running into a lot of problems. My latest is after tracking down a massive air leak I ran a tune - all looked good, not great. Then all of a sudden my peaks went haywire.
I finally figured out no PFTBA in the vial. I refilled and then find out they just filled it 6 months ago - my old lab I never needed to fill the vial in over 8 years. This is a 5975 so the vial is horizontal instead of vertical and when I removed the vial since I have only done this once I couldn't remember if there is supposed to be an o-ring around the collar.
Sorry lengthy post - possibly due to major frustration about lack of documentation and basic maintenance. :evil:

Re: pftba being used up within 6 months

Posted: Wed Jul 23, 2014 10:02 pm
by James_Ball
I have started in a new laboratory and they had a lot of preventative maintenance not being performed so I have been running into a lot of problems. My latest is after tracking down a massive air leak I ran a tune - all looked good, not great. Then all of a sudden my peaks went haywire.
I finally figured out no PFTBA in the vial. I refilled and then find out they just filled it 6 months ago - my old lab I never needed to fill the vial in over 8 years. This is a 5975 so the vial is horizontal instead of vertical and when I removed the vial since I have only done this once I couldn't remember if there is supposed to be an o-ring around the collar.
Sorry lengthy post - possibly due to major frustration about lack of documentation and basic maintenance. :evil:
I know how you feel, sometimes trying to keep people interested in doing maintenance is a full time job. Had someone last week ask me to look at their 5973 because it wasn't tuning and had air at about 10%. I looked at the autotune they just ran and the EMV was set to 3000V. It was the original EM that was in the instrument when it was installed in 2005!

As for the PFTBA, I believe it does go a little faster in the 5975 because of the way the vial is made. You just can't get as much in there without it overfilling. If there is an o-ring on the vial it should be inside the fitting, either an o-ring or a Teflon ferrule should be there to seal it.

Be sure not to fill it above the horizontal tube or you will be drawing in the liquid instead of the gas from the headspace. Fill a little then turn it sideways to make sure it is not too much.

Re: pftba being used up within 6 months

Posted: Thu Jul 24, 2014 11:47 am
by chloesabrina
Thanks James_Ball, it seems to finally be looking a lot better. I had also installed an oil free pump and so I checked all o-rings and added a little grease to help seal.

Re: pftba being used up within 6 months

Posted: Thu Jul 24, 2014 1:53 pm
by James_Ball
Thanks James_Ball, it seems to finally be looking a lot better. I had also installed an oil free pump and so I checked all o-rings and added a little grease to help seal.
Had they been running an autotune every day to use it up so quickly? I know some people do that as an ongoing check, but I just have always used my daily EPA tune checks so I hardly ever have to open the pftba valve.

I have wondered how well those oil free pumps work, I hope they have the lifespan that regular pumps have.

Re: pftba being used up within 6 months

Posted: Thu Jul 24, 2014 1:57 pm
by chloesabrina
Not sure how the oil free pumps work yet, will actually doing my first run today since installing it. They do get very HOT, but are quiet. They don't do too much EPA work here so they run a tune weekly. You think that could use up the PFTBA more quickly? I will have to make a note to check it more frequently, maybe when I clean the source just top it up.

Re: pftba being used up within 6 months

Posted: Fri Jul 25, 2014 7:58 pm
by James_Ball
Seems and Agilent Repair Engineer once told me that the 5975 would go through the pftba faster, since it actually did hold less in the vial than the older ones did. If they are running a tune ever week, it may be a good idea to check it monthly until you get a feel how long it lasts.