Washing out buffers at the end of the day...

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
I've been working with organic buffers, primarily 40mM ammonium formate/formic acid solutions. As is the case when I typically use buffers, at the end of the day I'll always spend the 30min to wash out my system with pure H2O, and the next morning I'll spend another 30min-1hr re-equilibrating my system to the previous conditions - but in this case I'm not exactly sure why I do it. Even at 95:5 MeOH:buffer, I know there is no chance for the buffer to precipitate as ammonium formate is soluble in methanol. Microbial growth might be an issue, but there is no circumstance when there isn't at least 30% MeOH running through the system.

For situations like this, is there really any necessity for me to go through this washing/re-equilibrating process everyday? What's going to happen if I leave formate buffer in my system/column for a day, a week, a month? Can anyone shed some insight on my situation? Thanks!
For one day I would set up flow rate at 0,1ml/min, that will save you 1 hour washing time. For 1 week or one month I would rinse the column and according to the manufacturers manual store in the recommended storing solvent. What ever could happen on the column with your buffer cannot happen then.
Gerhard Kratz, Kratz_Gerhard@web.de
We do what Gerhard has suggested, as it saves both flush time and reequilibration time for next-day use.
Meaning you just allow the system to run at a 0.1ml/min flowrate overnight, obviously with the detector lamp off, to keep things flowing? Thanks.

I guess I should just install the solvent valve module, that way switching between solvents won't be such a pain.
Meaning you just allow the system to run at a 0.1ml/min flowrate overnight, obviously with the detector lamp off, to keep things flowing? Thanks.
Yes.
Do you see an improvement now?
Gerhard Kratz, Kratz_Gerhard@web.de
6 posts Page 1 of 1

Who is online

In total there are 20 users online :: 3 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Google [Bot], John Guajardo and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry