GC Temperature error.

Discussions about GC and other "gas phase" separation techniques.

7 posts Page 1 of 1
Hello guys,

Today we stumbled upon a very strange problem with our Chrompack CP9001. The temperature settings for the GC were 70 degrees for the oven, 250 for the injector and 250 for the detector.

The oven won't reach the 70 degrees however, it stays stuck on 96/95 degrees. But when I measure the temperature of the oven it is 29/30 degrees. So somehow the displayed temperature on the GC is way too high. Also the temperature isn't stable. The detector and injector temperature vary between 247 and 253 (normally it stays perfectly at 250).

Also the temperature setting for the Aux1 and Aux 2 is 0, but the actual temperature is 44 (normally it also says 0).

I think that the GC somehow displays all the temperatures too high (higher than reality) and I have no idea how this is possible.

I recently uninstalled the FID detector and installed a FPD detector (but yesterday the temperature was still OK with the FPD detector and it happened overnight).

Any help will be appreciated.
Chemical Engineer at Instral BV.

Your specialist in solutions for:
  • Oilfield chemicals (desulfurization)
  • Water treatment (waste water, cooling water)
  • Mining solutions (anti dust, dust control)

For info:
http://www.instral.com
or:

r.wentink@instral.com
Sooting from the hip here: Maybe leaking/bad condensators on the board?

Condensators go bad over time, so if it is an old instrument it may be worth checking out.
Hello and thank you for your response.

I have looked into the diagrams of the GC and I see that there are a lot of condensators. How do I know which one I should check? (and how can I see whether a condensator is 'bad' or in good shape?)

Thanks
Chemical Engineer at Instral BV.

Your specialist in solutions for:
  • Oilfield chemicals (desulfurization)
  • Water treatment (waste water, cooling water)
  • Mining solutions (anti dust, dust control)

For info:
http://www.instral.com
or:

r.wentink@instral.com
Hello and thanks for the efforts (so far :P )

I have checked al the capacitors and all I could find was this:

Image

On the image the C25 and the capacitor under the IC38 block look like they might have leaked or is this negligible? I hope one of these are causing the problem because our whole GC system is useless at the moment.
Chemical Engineer at Instral BV.

Your specialist in solutions for:
  • Oilfield chemicals (desulfurization)
  • Water treatment (waste water, cooling water)
  • Mining solutions (anti dust, dust control)

For info:
http://www.instral.com
or:

r.wentink@instral.com
I dont know if they could cause your problems. Are they conected to the temperature sensing circuts?
C25 looks to me like a paper foil capacitor not a liquid filled electrolitic.
Board swapping is so much quicker and easier although not easy with a 20+ year old GC that did not sell in huge numbers.
Sometimes it is time to put old GCs down :wink:
Do the maths, how much for a working GC second hand/new vs the time and effort to fix an old one.
If you are in a commercial environment it must be time to move on ?
7 posts Page 1 of 1

Who is online

In total there are 141 users online :: 0 registered, 0 hidden and 141 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 141 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry