are we not violating the 1st law of thermodynamics?
measured recovery = 120% !, was expecting less than 100%. any ideas?
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Typically, solution volumes are engineered so that the internal standard peak in both the reference and in the sample will be the same size. And almost all use the same concentrated internal standard solution to make the Rf mixture and to add volumetrically to the samples, so its actual concentration cancels out. If your IS area differs and it shouldn't, that's an issue.thanks for replying Mr. CSG;however, there is an internal std used in this assay--and it turns out that it's peak area count is less in the test sample, than in the reference std., but shouldn't both the query analyte AND the IS retain equivalently?