hi, I have been using varian 320 uv/vis detector for two years ,yet ,gradully I find that ,the up limit of the absorb value often changs greatly when I change the wave length,and sometims the up limit changes even though I did northing with the detector,i really do not know why is so ?
Hi

I hope I can help you, there are several possible issues:

First the 320 Detector is running a calibration procedure if you switch it on.
In order to have a good calibration, it is recommended to rinse the flow cell 5 minutes before switching the 320 on ( just start the pumps a little earlier)
This should remove air or sample in the flowcell that might effect the calibration.

Second: There might be some dirt in the flowcell, we did talk about this before, espacially the dual-path cells ( like 9-1 or 4,5-0,15) are sensitive to this issue.
If you have a dual cell, there should be a x on the display

Third: The autorange function might be a possibility, too. You can switch it off
in "setup" --> "integrator/recorder"

Regards Chris
2 posts Page 1 of 1

Who is online

In total there are 48 users online :: 1 registered, 0 hidden and 47 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 47 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry